Results will depend on the type of equipment used in the analysis.
An Inductively Coupled Plasma (ICP) spectrometer will have an upper particle size detection limit of between .5 and 2 microns, and an instrument accuracy of .1 ppm. However, an ICP spectrometer requires the sample to be diluted and prepared prior to analysis. This can introduce sample to sample error much greater than the instrument accuracy.
A Rotating Disk Emission Spectrometer (Rotrode) will have an upper particle size detection limit of between 3 and 10 microns, and a measurement accuracy of 1 to 2 ppm. However, no dilution is required for a Rotrode analysis.
Differences between labs can occur for ICP in the preparation and calibration of the samples. Rotrode analysis does not have this problem, and will have fairly high repeatability between samples. But, Rotrode has lower accuracy.
Finally, since the upper particle size detection limit is higher for Rotrode analysis, reported ppm levels will generally be higher than ICP measurements, since particles of larger size are included. This will often explain the difference in a Blackstone and a Dyson analysis, since Blackstone uses ICP and Dyson uses a Rotrode lab.