Regarding SPEED:
"Because we use the Rotating Disc Method (RDE) for the atomic emission spectroscopy, it provides different results than the more common ICP (Inductively Coupled Plasma).
Because RDE doesn’t require the sample to be diluted, it can see larger particles up to 10 microns. On the other hand, ICP requires the sample to be diluted in a solvent, which lowers the maximum particle size that can be detected down to around 5 microns.
This change in particle size detection has a direct impact upon the parts per million of Molybdenum that are detected. The MoDTC additives read higher with RDE.
Simply put, you can’t directly compare the additive levels from the same sample between ICP and RDE. There will always be differences. If you know the differences, then you can adjust your perspective accordingly.
That’s a long way of saying that the calibration is not off. The equipment at the lab is calibrated every day."
"Because we use the Rotating Disc Method (RDE) for the atomic emission spectroscopy, it provides different results than the more common ICP (Inductively Coupled Plasma).
Because RDE doesn’t require the sample to be diluted, it can see larger particles up to 10 microns. On the other hand, ICP requires the sample to be diluted in a solvent, which lowers the maximum particle size that can be detected down to around 5 microns.
This change in particle size detection has a direct impact upon the parts per million of Molybdenum that are detected. The MoDTC additives read higher with RDE.
Simply put, you can’t directly compare the additive levels from the same sample between ICP and RDE. There will always be differences. If you know the differences, then you can adjust your perspective accordingly.
That’s a long way of saying that the calibration is not off. The equipment at the lab is calibrated every day."